Digital Multimeters (DMMs) and Precision DC Sources
Calculating PXIe-4145 SMU Resolution
I am trying to calculate the resolution of my PXIe-4145 4CH Precision SMU, but Figure 2 in the PXIe-4145 Spec Sheet appears to have inconsistent labels on the x-axis. If all labels were in msec (not mixed usec/msec), it would map to a consistent logarithmic scale as the graph implies.
aperture (sec) resolution (ppm) notes
0.010000000 1 if graph didn't curve
0.001000000 10
0.000100000 100 incorrectly labeled as 0.1 usec
0.000010000 1000 incorrectly labeled as 0.01 usec
0.000001667 6000 I think this is the max sample rate of 600kS/sec
0.000001000 10000 if graph continued
link to spec sheet:
https://www.ni.com/docs/en-US/bundle/pxie-4145-specs/page/specs.html#GUID-D04476BC-55C8-4CB8-9876-A9...
(5,460 Views)
Accepted by topic author sam.broyles
Yes, you're correct; the X-axis labels should be 1µ, 10µ, 100µ, 1m, 10m, and 100m.  Thanks for pointing this out.
(5,399 Views)
Hey Sam,
Thank you for bringing this to our attention.
I've filed a Corrective Action Request (CAR) to fix this in the 4145 specs (and also the 4144) specs.
Cheers,
Chris J.
Product Support Engineer
NI