Andrew,
Be aware that a DMM can only take one measurement at a time. The 4070 and 4065 can both sample very quickly, but since there is only one channel simultaneous sampling is not possible. Therefore to measure power you would either need to switch between the two measurments with a DMM (not simultaneous, but may work if your power is fairly constant) or use 2 DMMs. By using 2 DMMs you could send an external sample trigger to both DMMs so that they would acquire a voltage and current sample at the same time. Then you could calculate the power based on the two.
Have a good day,
Brian P., DMM Product Support Engineer
07-18-2007 12:23 PM
Andrew,
I also wanted to point out that the 4065 and 4071 can measure up to 3A without using any external shunt modules. You can go up to 10A with the external shunts that Jon linked to above. Below are a couple of other links that discuss power measurements with a DMM.
You could also use a DAQ board to measure voltage and current (using a shunt), but your accuracy and resolution would be worse than if you were using a DMM. You may also have to attenuate your signal in some way before feeding it into a DAQ board since they usually have smaller input voltage ranges.
Brian P, DMM Product Support Engineer
07-19-2007 10:26 AM
Thanks to both of you for the information.
I have decided to use PCI-4070 along with SCXI-1127 for my measurements
Thanks again