08-07-2003 03:42 PM
08-08-2003
12:18 PM
- last edited on
07-17-2025
09:00 AM
by
Content Cleaner
Hello,
There are not any Self-Calibration function calls for the NI 4060 DMM. The NI 4070 DMM is the only DMM that supports autocalibration at this moment.
You could use Calibration Executive to calibrate the 4060 boards. Calibration Executive 2.0 provides complete verification and adjustment of the 4060 product line. You will need to use an external instrument, the Fluke 7500 specifically, to perform your own external calibration of the NI-4060.
See also the links below for more information:
1)Calibration Executive
2)Calibrating NI 4050/4060 DMMs with Calibration Executive User Guide
2)Calibrating NI 4050/4060 DMMs with Calibration Executive
The final option is to send your board to us so we can Calibrate it for you. See more information in this link: https://www.ni.com/en/shop/services/hardware/calibration-services.html
Best regards,
Claudia Lorente
Applications Engineer
National Instruments
08-08-2003 12:19 PM