Digital Multimeters (DMMs) and Precision DC Sources
NI USB-4065 External Trigger slow sample rate
I am measuring voltage with the above device, using mesaurement complete and an external trigger.
The resolution is set to 3.5 digits, but I am getting a reading approximately every 100ms. The manual
states that at this resolution, the sample rate should be 3000/s. How can I get this faster rate?
Thanks.
(6,425 Views)
Could you supply some code?  The 3kS/s is when in waveform capture mode.  So I am not surprised by you not reaching that.  But the 100ms makes me think the DMM is doing some configuring.  Did you turn autoscale off?
(6,420 Views)
Hi crossrulz,
Thanks for the reply. I can't find any reference to auto scale, only autorange (which is only used for immediate sampling).
I have modified my code (I made a slight error) and am now getting 25ms per read.
You mentioned that the 3k/s is for waveform capture, do you know what sort of sample rate I should be getting?
Much appreciated
(6,396 Views)
OPullin wrote:  I can't find any reference to auto scale, only autorange (which is only used for immediate sampling).
Autorange is what I meant.  I must have had oscilloscopes on the brain when I typed that.
OPullin wrote:  do you know what sort of sample rate I should be getting?
Not a clue.  But as long as you are getting the rate you want, that is what matters.
(6,392 Views)
Configure the device for multi-point acquisition and you should get more measurements per second.
Marcos Kirsch
Chief Software Engineer
NI Driver Software
(6,381 Views)
Hi all,
Thanks for the reply. I managed to improve the sample rate massively by reducing the settle time.
Cheers
(6,363 Views)