curios,
In your last post, you indicated that you need to measure up to 150 V and up to 650 Amps.
With our PXI-4070 DMM you can measure up to 300 V (both DC and AC). If you need a switch to work with the 4070, so that you can multiplex a signal into the 4070, then the SCXI-1127, 1128, 1129, 1160 and 1161 would all be possibilities depending on the what other specs you are trying to meet.
However, your second spec, 650 Amps, is much too large for any of National Instruments hardware and switches to be used with. To put it in perspective, the 4070 can only measure up to 1 Amp, so the amperages that you are looking to measure are much too large for one of our DMMs or DAQ boards.
I hope that this helps to clarify what National Instruments has available for your application.
Regards,
Jed R.
Applications Engineer
National Instruments