curios,
After looking over your last two posts, I have a few more pieces of information for you. First off, in response to your last post, where you asked:
@curios wrote:
If the ranges of voltage need to be measured is between 0 to 130VAC and 0 to 50VDC,
the ranges of current need to be measured stay within 0 to 70Amp AC and 0 to 600Amp DC. Do I still need to connect voltages and current to be measured to the respective transducers (or the voltage sensor / current sensor) and the have the signal pass through the SCXI analog input (for isolation)?
There is a question if the signals stay within the maximum and minimum ranges of dmm, could I still connect it to the transducer in the first place?
Any signal that falls within the ranges of the DMM's capabilities could be measured directly with the DMM. You would not have to run a signal through SCXI if it falls within the DMM's specifications.
That being said, I want to reiterate, National Instruments does not have any hardware available to measure the type of amperage that you are posting about. 70 Amps AC and 600 Amps DC are too large to be measured by the hardare that National Instruments has available.
After looking over your list of needs, it seems likely that a PXI-4070 would be capable of measuring the types voltages and resistances that you are interested in, and when paired with an appropriate switch/switches, a DMM may be able to handle all the voltages and resistances you are interested in. (But not the current measurements).
Finally, I have a suggestion for you. After looking at your list of needs, I realized that the type of system you are trying to set up is very large and complex. Speccing out this type of system in a forum like this one, tends to be slow moving, as it is hard to clarify the finer details of the system. To finish specccing out this system, I feel that you would be best served by contacting your local NI Field Sales Engineer. Your local engineer would be able to come and meet with you and help you pick out appropriate hardware and software for your system. If you don't know your local engineer, call 1-800-433-3488 (in the US & Canada), and someone will be able to put you in contact with the appropriate person. If you are outside of the US/Canada, then please post your location, and I will find the appropriate contact information for you.
Regards,
Jed R.
Applications Engineer
National Instruments