Digital Multimeters (DMMs) and Precision DC Sources
Using PXI 4110 as precision DC current source and Measure Unit (SMU) )
   I want to use PXI 4110 as precision DC current souce and measure unit.Can it be used as SMU? I am using channel 2 (2mA to 11mA, Range :20cmA). Sourcing a current sweep of 2 - 11 mA with a step size of 1 mA through a resistor of 173 ohm, the readback voltage and readback current (readback current * 173) are not same.  I want to know its reason.
Khan97
(6,705 Views)
Hi Khan83
The PXI-4110 can be used to source current in the range you described. When taking the current measurement, does it vary from sweep to sweep? For example, at 6 mA the readback voltage is 1 V the first sweep and .8 the second. Also, as we sweep through, is the error seen a linear error? Does the resistor heat up causing a change in the actual resistance?
Jason D
(6,674 Views)
No, the error is not linear, It increases upto 5 - 6 mA and then it again decreases. Resistor 173 ohm is at room temperature.
(6,641 Views)
Hey Khan,
Can you post a circuit diagram and a graph of your output data?
-John Sullivan
Problem Solver
(6,625 Views)