07-31-2009 08:43 AM
Hi,
have a 5070 dmm in a pxi rack, and have a problem
The DMM is init
DMM is configured for several 2 wire resistance tests, no problems
Mode is changed to DIODE
a Diode measurement is done, no problems
Mode is changed to DCV, autorange
First voltage measured is always wrong: should be near zero, get about 10vdc consistantly
Second and all other voltages measure correctly
Anyone seen this behaviour before, I have use the 5070 a lot in the past, have never seen this
Jeff
08-03-2009 12:10 PM
Hello Jeff,
Similiar to a Resistance measurement, we provide a current through the DUT for a DIODE measurment. I am lead to believe that the device is still providing a current when you switch to a DCV measurement. Can you provide an image of your code, I am suspecting if you are switching these modes programmatically, you are not providing enough of a delay between measurement types?
Regards,
Glenn