4472B finite sampling

Solved!
Go to solution

Hi All,

 

I am using a 4472B to take a finite sampling with a digital falling edge reference trigger. The sample per channel is set to x and the pretrigger amount is 90%x. Now we start taking data using the StartTask (task_handle) in LabWindows/CVI. Suppose a falling edge arrives soon after the measurement task is started in such a way that there won't be enough time for 4472B to acquire 90%x of data points. In this case, is it possible that 4472B will neglect this first reference trigger and continue to fill up the pretrigger portion until the next reference trigger arrives?

 

Thanks!

 

XR

0 Kudos
Message 1 of 3
(6,621 Views)
Solution
Accepted by topic author magnetsu

Hi XR,

 

I believe I understand what you're asking, but let me just repeat it to clarify:  You'd like to know what happens when you set up a reference trigger for a DSA device after the task begins running in hardware, but enough time hasn't elapsed to capture all of the pretrigger samples you specified before the trigger arrives.  Is this correct?

 

I included a visual aid that hopefully will make things clearer, wherein 10 samples are acquired.  Nine (9) samples are specified as pretrigger sampled (90% of 10) and the remaining 1 sample is a post trigger sample.  In the image, the earliest the reference trigger can arrive in order to capture all 9 pretrigger samples is between samples 9 and 10.  This will arm the device to collect the remaining 1 sample (the 10th sample) as post-trigger.

 

The DSA Manual (on page 2-17, or p. 30 of the PDF file) explains what happens in this situation.  The device will ignore any triggers until first the pretrigger samples fill up the buffer allocated for those samples.  After the pretrigger samples are filled in that portion of the buffer, the pretrigger samples will continue to update with the most recent samples until the trigger is sensed, and the remaining portion of the buffer will be filled up with post-trigger samples.

 

Put another way, once the pretrigger samples are acquired, the circuitry then waits for a trigger to arm the remaining samples in the post-trigger data.  So the trigger circuitry will not affect the pretrigger or post-trigger functionality until all of the pretrigger samples are taken first.

 

I hope this helps answer your question.  Please let me know if I can help any further,

Regards,

Andrew

National Instruments
Message 2 of 3
(6,593 Views)

Hi Andrew,

 

Thanks a lot for your information and illustration! It explains the behavior of 4472B in my application. Thanks also for pointing out the discussion in the latest version of the manual (November 2010). I was looking at version August 2008...

 

Again, thanks a lot for your help.

 

XR

0 Kudos