Thanks for the reply, I'll gladly attempt to post some screenies.
This occurs with either AC or DC coupling, and at all ranges of sampling rates. The signal used here is 8V peak-to-peak, and the only change going from the sine wave to the noise input is the press of a button on the waveform generator. Again, the scope shows everything normally.
The auto-scaled noise sample shows a -150 mV average. This gets closer to 0 as the input voltage is taken down (as does the overall small amplitude of the signal).
The last plot show 2 channels from 3 cards, detailing the harmonic match, but failing the match the amplitudes of the diminished signal on each channel.
Eric
01-19-2008 11:42 AM
01-21-2008 02:16 PM
Ed,
That is certainly, in theory, what is occuring here. Let me follow up with more information to see if it fits.
I've tested two different noise generators. One in particular has a range of 20 Hz - 20 MHz. Another is a waveform generator with a max range of 15 MHz. I've also sampled at the max rate for these cards, 102.4 kS/s, while trying to figure this problem out.
What confuses me is that the 4472B card has two lowpass filters. The first is a two-pole analog Butterworth filter with a 400 kHz cutoff frequency which occurs before the signal reaches the ADC. The second is the 'brick-wall' digital filter tied to the sampling rate to remove aliasing. Testing the digital filter, it appears to be working as intended, as sine signals over 50 kHz bring the measured signal down to zero. So, assuming the analog filter is working, I have a range of frequency between 50-400 kHz being discarded or about 90% reduction in energy. Given a 8Vpp signal, I would expect measurements with a 1Vpp range... instead I'm getting less than half that, and a nasty aforementioned bias to go along with it.
I pulled out an old (and still very useful) spectrum analyzer which outputs noise with a max range of probably 25-50 kHz. Being within the nyquist rate, the corresponding acquisition was dead on. Not having any low-pass hardware filters on hand, this will serve our needs for performing channel phase calibration. Our actual experiments won't have data beyond the 5-10 kHz range beyond noise, and ultimately a perfect channel calibration signal would include just the experimental frequency range. I was just being lazy and using available equipment.
I'm not certain about the analog filter either. When ramping up the sine wave frequency from the waveform generator over the nyquist rate up to a maximum of 15 MHz, no signal existed, but a negative bias did exist, up to -0.5 V at around 6 MHz. This is coincidentally close to a narrow hole in the digital filter at 6.55 MHz for max Fs, but not exact. If the low-pass analog filter was doing it's job, why is there this bias from the high frequency components?
In previous experiments a couple years ago with other hardware, we sampled noise from these other noise generators without problem. I'm disappointed that apparently the ADC digital aliasing filter is being overloaded from signals having 90% frequency content over the nyquist rate. As I'm just coming on-board with setting up the equipment for signal analysis, I'm not being very knowledgeable about the breadth of hardware capabilites, so maybe I shouldn't be surprised.
Thanks for your knowledgeable input,