Hi Giovanni,
> thanks for yor answer. As a matter of fact I'd exclude any kind of
> ground loops and/or shielding problems. They're usually such a crucial
> issue (especially with pulsed excitation) that I took great care about
> them. As long as with a AC excited bridge I get a 'perfect' signal, I
> do think that with FP there's a combination of: transient response to
> step signal, an irregular self heating (although the excitation is
> pulsed, voltage is applied to the bridge for some tens ms), and an
> unstable excitation. Sometimes I had similar problems with FP RTD
> modules.
Same with me. In all the past I used AC- exited bridges and allways had
perfect signals. Maybe this is interesting for you. I connected a
oscilloscope to a channel of a SG-140, with the adjustments half-bridge
completition = on, Uexc = 5V, filter = 15 Hz, Range = +/- 3,9 mV/V. The
measured Ti = 320ms, Tp = 2250ms. I monitored this with the FP-Explorer in
default configuration- just monitorin one channel.
Vexc(Ti) = 4.7V, Vexc(Tp) = 0V; Vin+(Ti) = 2,4V. Instead of strain gauges,
I connected 2 metalfilm resistors with 120 Ohm/ 1% each as a half-bridge
to the module for testing. The signal shape itself was a perfect
pulse/pause (Ti/TP).
So I don't really think you have to worry about the shape. And the signal
voltage the FP-Explorer showed was very stable. It sometimes changed about
+/- 100nV. If you keep in mind that 16bit resolution in this input range
means a resolution of about 119nV/V, this is a very good result.
There is another thing to be aware of, if you use the SG-140 with
half-bridges. If you do so and swith on the half bridge completition, you
will notice that you won't get a 0-display if the SG-140 input is feed
with 0.000mV. The module- internal halfbridge completition has an offset
with value about +/- 1.xxx mV/V and is different for every channel and
module. I had this tested with 2 new out-of-the-box module, and one
borrowed one of NI. Maybe you've seen this a couple threads above. There
are no problems like this, if you use full- bridges.
Maybe you should give it a try, just connecting metalfilm resistors
instead of the load cell for testing purposes. If the unsteady signals
still show, the reason must be some place else, but not a matter of the
load-cell.
Good look,
and please keep me on track, what youre doing, because there aren't to
many people using the SG-140.