FieldPoint Family
Set Resistance for cFP-RTD-122
Hello Forum,
I use a cFP-RTD-122 to measure temperatures of PT100. My problem is, that our sensors don't have exactly 100 Ohm but rather 115-120 Ohm. How can I set this value? The only thing I found is the possibility to change the TCR value for each channel which doesn't solve the problem.
Thanks for help,
Christoph
(5,351 Views)
Hi Christoph.  Are your sensors at 0C when they read 115-120 ohms?  PT100 sensors
should be 100 ohms at 0C and 115-120 at 40C to 55C.  How long are your lead wires, and
what gauge?  Are you using 3 wires to the sensors?
If your sensors are really PT100 sensors, then selecting that and the proper TCR should
give you a good reading.
If your sensors are something other than PT100, my guess is that you will have to setup
the RTD-122 to read resistance and translate to temperature in your code.
Matt
(5,350 Views)
Hej Matt,
thnaks for your fast answer.
I have here a thermal manikin which has measurement wires below the surface. The wires used are PT100 but the resistance changes a bit because of the used length an the connectros and cables attached to them. I use a 3-wire-measurement and the measured 115-120 Ohms are at 0C.
I guess I have to calculate the temperature from the resistance then?
/Christoph
(5,348 Views)
Wow.  15-20 ohms is a lot of lead resistance.  Any chance you can reduce
that with bigger wires/fewer connectors?
What temperature does the RTD-122 give you if you just connect the sensors and
set to PT100 and the proper TCR?  You might get away with just a calibration offset,
i.e. RTD-122 reading - 20C = actual temperature, depending on the temperarure
range and accuracy required for your application.
Matt
(5,344 Views)
I can not change anything on the manikin without destroying it.
I tried the thing with the offset you mentioned as well. The problem is that it is not a constant offset but rather linear dependet on the temperature.
What I do no is calculate the offset from a calibration  and subtract this from the measured offset.
Thanks for your help,
Christoph
(5,342 Views)
Sounds like you will have to take readings at several temperature points
in your range and do some interpolation/calculations.  At least you are
aware of the problem.  Sorry I don't see any simpler answers.  Good luck
with it.
Matt
(5,339 Views)