 5122-user
			5122-user
			09-25-2014 08:09 PM
Hi all,
I had to replace a 5124 card with a 5122 card in an existing system. nilsdev detected the new card but said
NI PCI-5122: "Dev1" [Failed to Load]
To test things out I have setup a new system using openSUSE 11.4, installed ni-scope 3.1 (and before that I set up nikal 2.3.1 manually because the older nikal version which comes with ni-scope 3.1 is not compatible with this kernel). It took me some trouble to install everything as the install script of ni-scope tries to verify the existence of the linux sources by searching at depreceated locations. However, I could fix all of this and successfully install the driver. But in the end I got the same "Failed to Load" message as on the old system.
I have attached the output of niSystemReport.
I would really appreciate any help to get our 5122 card running on our linux system.
Many thanks,
Frank
09-26-2014 10:46 AM
I decided to test the card on a Windows 7 machine. There, I get the following error when running the Self-Test in NI MAX:
The device self-test has failed. The error report from the device driver is as follows: Error -239022 occurred at Self-Test
Possible Reason(s): Internal Hardware Error occurred in MIG hardware. Please contact National Instruments Support.
09-26-2014 11:11 AM
Can you please provide your module's Serial Number and Part Number?
09-26-2014 11:40 AM
Sure:
Serial number: E50875
Part number: 190971C-01
Thank you!
09-26-2014 11:54 AM
Also I just noticed that on our other card, the PCI 5124, a green LED lights up on the board and the fan starts running as soon as the system is powered up. However, on the 5122, neither the green light turns on nor the fan starts running.
09-26-2014 01:10 PM
Confirmed, your failing board needs to be sent to NI for repair.
https://www.ni.com/contact-us/
Nathan