High-Speed Digitizers
Average on board of NI5922
Hi
Is it possible to sample at 15 MS/s and calculate the average of 15 Samples on board of NI5922? That would rusult in 1 MS/s averaged datapoints.
I use NIScope 3.8.1 and LabView 10.0.1 (64-bit) Software.
Thanks!
Tobi
Message 1 of 12
(7,785 Views)
Hi Tobi,
as you can read in http://digital.ni.com/public.nsf/allkb/657C83B20DF586678625720D0028A7D1?OpenDocument,
you can use Onboard Signal Processing (OSP) in some cases, but it is more like for doing filtering, decimation and related stuff.
"...For instance, all functions and properties labelled NI-SCOPE Measurements are done in software (Array Measurements, Waveform Measurements, and Scalar Measurements). This includes Windowing, Histograms, IIR and FIR digital filters, FFTs and Power Spectrums, Interpolation, Averaging, Max/Min Detection, etc."
So averaging won´t be done on the Scope´s FPGA, but that´s no issue, just use the respective NI-Scope-Funktion to average your samples.
regards
Marco Brauner NIG
Message 2 of 12
(7,754 Views)
Hi Marco
Thanks for your answer!!!
Actually I didn't´t find an appropriate averaging function in NIScope. There is "RIS Num Avg" in the NiScope Property Node, but I don´t think it´s the right function... Do you know a specific function?
At the moment I do it as followed:
- fetch digitizer data (e.g. 2000 points per fetch, at 2 MS/s)
- Send the digitizer data into a queue
- build the averages with the attached sub-vi after dequeueing in LabView
- write the averages to TDMS datafile
Problem: the averaging seems to be very slow and the bottleneck (my queue fills up!!!). Maybe the conversion back to I16 is problematic... How could I build averages straight away from I16?
any suggestions to make it faster or do it differently?
Tobi Broger
Institut für Mikrotechnik Mainz
Message 3 of 12
(7,742 Views)
Im also interested in this question since we use quite a bunch of them 😉
I haven't found an answer in the Manual,  KB.
The 5922 is a flex resolution card with a tricky delta-sigma converter. So the answer might be buried in the way the ADC works. Other scope cards have different converter architectures.
My current understanding is, that the card constantly sample with 120MHz 6bit delta-sigma and the chosen 'samplerate' determine the way of decimation/averaging/weighting/filter
Might be, that if you sample with 1MS/s the value is already a short term average of 1µs, but the aliasing filter diagram (fig 16 in the paper here or see spec) is an indicator that it is not.
And to extend the question: In lower samplerates, will there be a point where pure decimation is used?
Please NI point to some (or publish) papers that put light to that question.
Maybe a solution is to use the on board FIR filter to do the averaging (read the filter, fold it with a rectangular window (your averaging), write back) and use a simple decimate. BTW have you tried the decimate (continous).vi ?
Greetings from Germany
Henrik
LV since v3.1
Message 4 of 12
(7,730 Views)
Hallo Henrik
Thanks for your interesting post!
If it is really the case that i would already get the averages at lower sampling rates I wouldn't have to do anything at all. I will perform the following experiment to find out: measure noise at 1 MS s-1 and at 10 MS s-1: If the Amplitude of the Noise is similar >> there is no averaging, If the amplitude of the noise at 1 MS s-1 is lower >> there is averaging (Averaging shoud reduce noise, right?). Does that experiment make sense?
I will also have a look at decimate.vi and get back to you asap!
Greetings from a Appenzeller in Germany
(Which is quite unusual at the moment, most Swiss do only shopping in Germany at the moment 🙂
Tobi
Message 5 of 12
(7,726 Views)
Already did the experiment
I fetched data (I16) at different sampling rates and measured the Amplitude (i was to lazy to scale it to Volts, as it is not important for the comparison):
Results:
1 MS s-1 >>   Amplitude 290
10 MS s-1 >> Amplitude 1170
10 MS s-1 and with AverageSignal.vi >> Amplitude 300
It really seems, that NI5922 does averaging on board by default!
Could someone from NI approve this?
@Henrik: I had a look at decimate.vi. It would be perfect to use - but i guess it won´t be necessary anymore ;-).
It would be great to hear your comment about the little experiment above!
Ciao Tobi
Message 6 of 12
(7,722 Views)
The expected noise reduction would be SQRT(10) .. about 3.16   you measured factor 4  ... what type of noise did you use (had a look at the spectrum)?
Henrik
LV since v3.1
Message 7 of 12
(7,710 Views)
Accepted by wickiwau
If I remember correctly (and it has been awhile), the 5922 has a 6-bit ADC and gets it resolution by averaging.  So if you take a 1MHz sample rate, you should get the time average of the signal you would get if you averaged ten points taken at 10MHz.  This is not true for other National Instuments high-speed digitizers (non-flex-resolution).
If you prefer to use an explicit averaging, I rewrote your code and got about an order of magnitude speed increase.  You can get more by turning off debugging in the VI properties.  You can also set it to subVI priority, but only do that if you can guarantee you will not be calling it simultaneously, or it will slow you down a lot.
Let us know if you have further questions.
Message 8 of 12
(7,708 Views)
Maybe it's my lag of understanding deltasigma converter , but this card is special ... and some kind of workhorse here, so everything bit of extra information is wellcome.
(It took me long time to understand the HP3458 in detail (much more than a dual slope converter 😉 ) but you can get about 4-5kg documentation for it.
Henrik
LV since v3.1
Message 9 of 12
(7,702 Views)
Thanks a lot for the constructive discussion and the code! Very kind! I´ll prefer methode1: the "averaging" on a lower sampling rate!
Best wishes
Tobias Broger
Message 10 of 12
(7,694 Views)
