Hi Darrow,
First I will link to the Manual Calibration Procedure and Specifications for this board for reference.
There are a couple of sources of noise on this module that need to be taken into account. One shown in the vertical sensitivity table which you have already referenced. This is amplifier noise density which is dependent on gain settings. The second is ADC noise density. This is dependent on sampling rate. These values can be found in the Dynamic Range section of the Specifications.
The two noise sources are uncorrelated, and depending on your combination of sample rate and range, one will typically dominate and can be used as an estimate. Otherwise you can obtain the RMS total through the root sum of squares method. This total is what is used in the calibration tests.
There is a VI that has been set up for calculating the noise-floor of the 5911. You can find it here if you wish to see an illustration of the calculation (check out the block diagram).
If your module is failing calibration for any reason, you cannot rely on it meeting the accuracy specifications. This specific type of failure will impact the dynamic range/noise floor of the board. Likely tens of dB that you have lost given the measurements you report. Small signal measurements will definitely be affected. I recommend repair or replacement to ensure the quality of future measurements.
Jennifer O.
06-27-2008 02:43 PM
07-08-2008 10:50 AM
Given the readings in my original post (elevated noise density at 0.2V range and low sample rates) and the gain noise and ADC noise specifications of the 5911, would it be reasonable to assume the source of the elevated noise density in my 5911 was probably a component in the amplifier circuit?
Thanks,
Darrow
07-08-2008 12:50 PM
07-08-2008 04:31 PM
Hi Jennifer,
I'm evaluating the impact if any on measurements made prior to the repair. The digitizer was only used in the +/-5V range at 12.5MS/s sample rate to measure a 2 square wave during that time Given the noise specs of the 5911 relative to range and sample rate settings, compared to the call exec readings noted in my original post it would appear the source of the elevated noise was the amplifier. The cal exec readings in the 20V range at high sample rates where ADC noise would dominate were were not effected. I received a repair report from NI however it simply states a component was replaced with no other information.
Darrow
07-10-2008 01:01 PM
07-10-2008 02:19 PM