High-Speed Digitizers

cancel
Showing results for 
Search instead for 
Did you mean: 

In PCI -5154 I want to acquire only peak of the sine wave.How to do it.

Hi ,

I have PCI-5154 digitizer.I want to acquire only single point in a wave , say in sine waveform i want to acquire only peak not all data.How can i do it.And also What is the memory 256MB on board memory mean?

0 Kudos
Message 1 of 4
(6,336 Views)

Digg into the manual of NI-Scope and you find

scope-scalar-meas.png

the scalar measurements provide a list of supported on board measurments including Peak-to-Peak-Voltage ...

 

 

(or in short RTFM SCNR 😉 )

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 2 of 4
(6,322 Views)

I didnot understand it.Actually i dont want acquire and analyze of wave .I want to acquire only peak of the sine wave.Which way is suitable whether triggering ,timer.

0 Kudos
Message 3 of 4
(6,315 Views)

It's hard to make a valid measurement if you don't understand it 😉

Set up a ni-scope task and configure the samplerate at least n=20 times the measured frequency with at least 2n samples , then instead of reading the waveform and analyze it by your own,  use the vi noted and you get the peak to peak value. Divide it by 2 and you have the amplitude 🙂  (and you compensated a possible offset at the same time 😉 )

 

Are you shure that your (frequency stable!) sine is in sync with your sampleclock?

 

Maybe it's even more simple to aquire when you use the express-vis, but I never used them for serious tasks, since then I want to know what I do...

 

 

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 4 of 4
(6,311 Views)