01-31-2019 02:48 PM - edited 01-31-2019 02:49 PM
I can export AnalogComparisonEvent to any terminal like PFI0. This signal is supposed to be a TTL with max = +5V.
But it seems the impedance of the PFI is not 50Ohm, because when I look the signal on a scope (terminated 50Ohm), the maximum is at only 2.4V.
Can I change this impedance ? How can I manage otherwise to get a +5V TTL on a (50 Ohm load) out of my card ?
Config. : a PCIe 6259 (M-serie) or a PCI 6110 (S-serie), with BNC-2110 or BNC2120.
Thanks !
02-13-2019 01:09 AM
external hardware buffer/driver , powered by the provided 5V
02-13-2019 10:14 AM
Thanks. So no other option than to use an additional external device that will amplify then signal, that's it ?
02-13-2019 11:26 AM - edited 02-13-2019 11:27 AM
If you use more that one PFI and parallel them ... but I don't recommend that.
read the spec, all expected behaviour.. If you increase the load impedance you get closer to to 5V. 50Ohm is no typical TTL input 😉
What is the problem? 2.5V is fine to trigger something 😄