05-24-2013 04:50 PM
I am trying to read data with the NI-5772. I have tried the example code(NI 5772 - Getting Started (Host).vi) , and it will read noise (with nothing plugged in to either channel). However when I attempt to route a signal in (such as one generated with a NI-DAQ board on the same system and routed to AI0/AI1), the signal doesn't change much from the noise. I updated the drivers and am running LabView 2012 (both 32-bit and 64-bit, I get the same result)
Has anyone else experience this problem? Thanks.
Solved! Go to Solution.
05-24-2013 06:07 PM - edited 05-24-2013 06:12 PM
Ok, apparently it just didn't like our input signal ( a 1 khz sine wave generated from NI-MAX). We tried a source with a higher output rate (several MHz) it seems to work OK. Is there some sort of high pass filter on the input?
05-28-2013 09:28 AM - edited 05-28-2013 09:31 AM
Yes, the device is AC coupled and has a high pass filter. There is a frequency response graph on p.20 of the User's Manual:
http://www.ni.com/pdf/manuals/373625a.pdf
It appears that the cuttoff frequency is about 100kHz.