2 gpib controllers on 2 different computers controlling one bus (not at the same time)

Solved!
Go to solution

Is there a way to achieve this whithout any conflict between the gpib boards ?

 

How to configure these gpib boards ?

 

Thanks

0 Kudos
Message 1 of 5
(4,454 Views)

Only one at a time can be configured as master or controller in charge (CIC). There are functions to transfer CIC and you will need to write your programs to be aware of each other.

0 Kudos
Message 2 of 5
(4,451 Views)

Thanks for you answer, before dealing with the programs, do I have to configure anything in NI-MAX on each computers before ? or the code I will write with the basic gpib functions will replace what NI-MAX does ?

 

 

0 Kudos
Message 3 of 5
(4,448 Views)
Solution
Accepted by Seb22

You'll have to use MAX at least once to make only one CIC. Otherwise you will get an error the first time you try to talk to an instrument. After that, your programs will take care of switching back and forth.

0 Kudos
Message 4 of 5
(4,445 Views)

I fail to see the worth of that control strategy and complexity. Sounds like a bad system design that could use some rethinking.