34401A sample rate problem

Hi,

I want to set the 34401A to high sample rate, but failed. I set sample count  to 5000, but the result always have 10 samples. And the Sample interval, it seems do not work.

Thanks!

0 Kudos
Message 1 of 5
(4,144 Views)

The 34401A doesn't support does not support Sample Interval or Pretrigger Samples.

Regards
Ray Farmer
0 Kudos
Message 2 of 5
(4,140 Views)

Ray is correct. If you open up the Configure Multipoint VI then you will see that this is noted on the block diagram:

 

sampleinterval34401a.jpg

 

If you look up the manual for the 34401A you will also find this documented there.

 

Lars L

NI R&D

0 Kudos
Message 3 of 5
(4,132 Views)

Hi Lars.

 

When a function is not supported it should generate an error when trying to set it.

 

As this is a NI suppported driver, please modify the driver so that users notice this.

A warning is not sufficient. If somebody wants to ignore it Ok, but no hiding of errors please.

 

 

greetings from the Netherlands
0 Kudos
Message 4 of 5
(4,129 Views)

Hey Albert,

You are correct. There should be a notification of some kind to alert the user that these functions are not supported for this particular model. I'm filing a bug report to get this implemented.

 

Lars L

NI R&D

Message 5 of 5
(4,117 Views)