Hello All,

I am using Agilent MSO8104A series Oscilloscope and I want to make a measurement (ie : Period) I use the VI exemple : "Agilent 8000 Series Histogram Measurement.vi"

but I have a Error : -113,"Undefined header"

it seems that the configure Measurement Histogram.vi is wrong

How can I have my measurement ? 

 

 

Software use : Labview 8.6.1

0 Kudos
Message 1 of 2
(3,514 Views)

Hello Rock,

       First, please take a NI-Spy capture of the example running with all of the default values.  For instructions on how to take a NI-Spy capture, see here.  Post the .spy to this thread.

 

Second, try changing the settings of the example and see if the error goes away.  Let us know if that helps or if it does not.

 

Third, you have the option of looking at the NI-Spy capture and finding out which command is causing the error.  You can then look that command up in the instrument's programming manual and see if the instrument driver is sending it incorrectly.  If it is, you can fix the command and see if that resolves the error.  After doing that, you can post the solution to this thread and the instrument driver can be fixed and re-released.

 

Cheers,

 

NathanT

Message 2 of 2
(3,487 Views)