GPIB Remote Command for R&S FSL Spectrum Analyzer to set "RF Atten Low Distortion"
Hi Juergen
After presetting the R&S FSL, I enter the sequence made before, and get the same softkeys menu as you shown on your picture, but with the softkey "RF Atten Low Distortion" under the "Input 50ohm/75ohm" softkey Smiley Happy
Please see my attached photo taken today.
I already contact R&S and they confirmed my situattion, and the missing remote command from the manual.
They forwarded my question to other department Smiley Happy
Message 11 of 14
(2,755 Views)
first of all I'd like to introduce you the R&SCommander tool: easyiest way to make screencopies of R&S gizmos: http://www2.rohde-schwarz.com/en/service_and_support/Downloads/Application_Notes?letter=Z&discontinu...
Same can be done by remote applications: http://www2.rohde-schwarz.com/en/service_and_support/Downloads/Application_Notes/?query=hardcopy&que...
The R&S Customer support is takeing care of that.
Regard juergen.
Message 12 of 14
(2,725 Views)
Thank you  Juergen!
Smiley Happy I made a screenshot on the FSL but got only the grid part. Thanks for the reference to the Commander Tool!
Regarding the low attenuation problem, I'll wait until one solution from R&S.
João
Message 13 of 14
(2,721 Views)
Accepted by topic author JPLO
Hi all
R&S gave the solution. The GPIB Remote Command for R&S FSL Spectrum Analyzer to set "RF Atten Low Distortion" is:
INPut:ATTenuation:AUTO:MODE NORMAL | LDISTORTION
Thank you Juergen for your patience Smiley Happy
Message 14 of 14