Hi Juergen

After presetting the R&S FSL, I enter the sequence made before, and get the same softkeys menu as you shown on your picture, but with the softkey "RF Atten Low Distortion" under the "Input 50ohm/75ohm" softkey Smiley Happy

 

Please see my attached photo taken today.

 

I already contact R&S and they confirmed my situattion, and the missing remote command from the manual. 

They forwarded my question to other department Smiley Happy

0 Kudos
Message 11 of 14
(2,251 Views)

Hi,

 

first of all I'd like to introduce you the R&SCommander tool: easyiest way to make screencopies of R&S gizmos: http://www2.rohde-schwarz.com/en/service_and_support/Downloads/Application_Notes?letter=Z&discontinu...

 

Same can be done by remote applications: http://www2.rohde-schwarz.com/en/service_and_support/Downloads/Application_Notes/?query=hardcopy&que...

 

 

The R&S Customer support is takeing care of that.

 

Regard juergen.

Message 12 of 14
(2,221 Views)

Thank you  Juergen!

Smiley Happy I made a screenshot on the FSL but got only the grid part. Thanks for the reference to the Commander Tool!

Regarding the low attenuation problem, I'll wait until one solution from R&S.

 

João

0 Kudos
Message 13 of 14
(2,217 Views)
Solution
Accepted by topic author JPLO

Hi all

R&S gave the solution. The GPIB Remote Command for R&S FSL Spectrum Analyzer to set "RF Atten Low Distortion" is:

 

INPut:ATTenuation:AUTO:MODE NORMAL | LDISTORTION

 

Thank you Juergen for your patience Smiley Happy