 ACar89
			ACar89
			05-15-2013 11:50 AM
Hello
On a remote system, I have HP 3458A and Lake Shore 331 Temperature control connected on GPIB. I have installed a NI PCI-6259 on the remote system, as well as NI VISA 4.2 and Real-time 11. Also, I have installed the instrument drivers of the two devices on my local computer.
Using VISAIC, I am able to see both devices however, I am only able to use the HP 3458A. The GPIB address of the LS 331 is incomplete; it is visa://[IP]/GPIB0::INTFC while the HP 3458A is visa://[IP]/GPIB0::22:INSTR
Any ideas?
Thanks in advance
Solved! Go to Solution.
05-16-2013 08:08 PM
You mentioned the LS is a GPIB device. Try installing NI 488.2 and see if it shows up.
05-17-2013 08:32 AM
Hello
I have installed NI 488.2 version 2.5 on the system. I just forgot to mention it.
Thanks though,
05-20-2013 08:20 AM
Have you tried to run a *idn command to the LS device? You also may try to run an example program in LabVIEW to see if you can communicate.
 JÞB
			JÞB
			05-20-2013 10:12 AM
It sounds like your seeing the controller and the DMM.
do you have the 331 set to use the GPIB? that must be done from the front panel.
05-23-2013 09:14 AM
I had tried running the *idn to the device and I am not sure what the 331 set is. However, it turned out that the GPIB cable going to the device I was trying to communicate with was damage (not exactly sure how) and it started communicating when I attempted to use a different GPIB cable.
All is good. Thank you for your time.