HP 70950a OSA
Hi
I want to extract data from my HP 70950a Optical Spectrum Analyzer (OSA). I'm not quite sure how to proceed with this, but what I found out (http://forums.ni.com/t5/Instrument-Control-GPIB-Serial/hp-70950a-osa-and-8163a-multimeter/td-p/11018... was that there are no drivers for this, so I need to do it "manually". In the VISA Interactive Control I have chosen GRIB0, and in the GRIB-control I have tried executing
OUTPUT\s723;"CENTERWL\s1300NM;SPANWL\s10NM;RL\s-5DBM;"\sSet\smeasurement\srange
as a test. It says "No error", but doesn't change the wavelength on the OSA. Does anybody have any experience with getting data from the device?
Best,
Karin.
Message 1 of 16
(5,480 Views)
Try updating the trace with the single sweep mode "SNGLS" or single sweep "TS" command or both.
and try the commands without "\s", just leave a space
Curt
Message 2 of 16
(5,476 Views)
Hi
Thanks for helping. It still says "No error", though, without doing anything. I'm not sure if I am doing it correctly, but I am Executing the commands in the GPIB contol of the VISA Interactive Control.
Best,
Karin.
Message 3 of 16
(5,473 Views)
Are you using an NI GPIB card?
Message 4 of 16
(5,466 Views)
Message 5 of 16
(5,464 Views)
Hi Karin,
Open up the NI-MAX application. It should be on your desktop. Select the GPIB0 card under devices and interfaces.
Click 'Scan for Instruments' and when your OSA shows up in the device list, click on it and then
click the 'Communicate with Instrument'
Click NI-Spy to start the spy app and when that starts goto menu Spy->Start Capture
Enter your commands, one at a time in the communicate window and see if you're OSA responds correctly
Hope this helps
Curt
Message 6 of 16
(5,461 Views)
Thanks, that's very kind of you. I'm not in the lab at the moment, but I'll try it our tomorrow, I'll make sure to check back!
Best,
Karin.
Message 7 of 16
(5,460 Views)
You aren't really including "OUTPUT\s723;"" in the  text of the interactive control, are you? That would be silly.
Message 8 of 16
(5,446 Views)
I did at first (silly, yes!), but I have also tried without.
Best,
Karin.
Message 9 of 16
(5,441 Views)
Thanks, now I can change the wavelength span and center by the above commands. However,
 I am still not sure how to do something like this (taken from the OSA manual):
10 CLEAR 723 Initialize analyzer.
20 OUTPUT 723;"IP;SNGLS;TS;"
30 OUTPUT 723;"CENTERWL 1300NM;SPANWL 10NM;RL -5DBM;" Set measurement range.
40 OUTPUT 723;"TS;" Sweep trace A.
50 OUTPUT 723;"MKPK HI;" Mark highest signal level.
60 OUTPUT 723;"MKWL?;" Return wavelength at marker.
70 ENTER 723;W Assign wavelength to computer variable.
80 OUTPUT 723;"MKA?;" Return amplitude at marker.
90 ENTER 723;A Assign amplitude to computer variable.
100 W=W/1.E-9 Convert wavelength from meters to nanometers.
110 PRINT W,A Print values.
Where can I write scripts like this, and execute them?
Thanks for all your help so far.
Best,
Karin.
Message 10 of 16
(5,411 Views)