I was seeing a weird bug in the Agilent 3446x driver on the driver page.
Running initialize.vi results in a long wait (~2 minutes) followed by this error:
Error -1073807339 occurred at VISA Read in Agilent 3446X Series.lvlib:Error Query.vi->Agilent 3446X Series.lvlib:Reset.vi->Agilent 3446X Series.lvlib:Initialize.vi
Possible reason(s):
VISA: (Hex 0xBFFF0015) Timeout expired before operation completed.
Then if I run the code again I get this error:
Error -1073807360 occurred at VISA Open in Agilent 3446X Series.lvlib:Initialize.vi->Ivan acquire single.vi
Possible reason(s):
VISA: (Hex 0xBFFF0000) Unknown system error (miscellaneous error).
I narrowed it down to a read call in Error Query.vi. I was able to fix it by inserting a 1 second delay in between.
I thought I'd write this up for anyone encountering this issue in the future.
I'm testing with a keysight 34465A firmware V, Windows 10, and Labview 2017.
07-06-2017 10:46 AM
Thanks for the info! I'm sure this will help people in the future.
You can also email instrument.drivers@ni.com with this write up to provide direct feedback.
01-08-2018 12:11 PM
Thank you so much, finally it works
This bug is still existing...
01-09-2018 09:18 AM
To echo Salazaar.Slytherin above, giving feedback to instrument.drivers@ni.com directly would be helpful feedback.
11-29-2018 06:48 PM
This is a sneaky bug in the driver. I was tinkering with the driver for 2 days. Thank you for the solution.
Added the delay in the error query.vi and it works perfectly
Windows 7 64bit with Labview 2017 32bits.
Now I can code!
06-12-2019 03:00 AM
Delay is needed once after the reset (speed up meas)