IEEE2 SRQ problem with METCAL

Solved!
Go to solution

I know this is Met/Cal related problem and I should ask Fluke Metrology community for this - but just in case someone here has a similar problem or solution...

Problem: all the new MET/CAL procedures using IEEE2 or SCPI commands result in SQR wait errors. Analysis using GPIB bus capture with National Instruments NI Spy software indicates that MET/CAL tests for TestSRQ(1, FALSE (0)) forever (about 300,000 times before giving up). Changing the SCPI commands to plain IEEE work just fine (they don't seem to expect SRQ from the device at all). The tested DUTs should be fully 488.2 compliant (eg. Agilent DSO8064 oscilloscope).

A test was carried out with the following set up:
MET/CAL 7.2.51, MET/CAL Editor 7.20Z
Procedure: Agilent DSO8064A: (1 yr) CAL VER /9500,1Hd,DMM
version: 2007-11-06 Revision: 1.2

Example command from procedure:
1.016  SCPI         *RST

Corresponding captured bus traffic:

1.  ibtmo(0, T10s (13))
Process ID: 0x00000F14         Thread ID: 0x00000624
Start Time: 10:33:12.625       Call Duration 00:00:00.000
ibsta: 0x170       iberr: 11             ibcntl: 0(0x0)

2.  Send(0, 0x0012, "*RST;*CLS;*SR...", 40 (0x28), DABend (0x02))
Process ID: 0x00000F14         Thread ID: 0x00000624
Start Time: 10:33:12.625       Call Duration 00:00:00.000
ibsta: 0x168       iberr: 0             ibcntl: 40(0x28)

3.  Send(0, 0x0012, "*OPC", 4 (0x4), DABend (0x02))
Process ID: 0x00000F14         Thread ID: 0x00000624
Start Time: 10:33:12.625       Call Duration 00:00:00.000
ibsta: 0x168       iberr: 0             ibcntl: 4(0x4)

4.  Send(0, 0x0012, "*ESR?", 5 (0x5), DABend (0x02))
Process ID: 0x00000F14         Thread ID: 0x00000624
Start Time: 10:33:12.625       Call Duration 00:00:00.000
ibsta: 0x168       iberr: 0             ibcntl: 5(0x5)

5.  Receive(0, 0x0012, "1.", 511 (0x1FF), STOPend (0x100))
Process ID: 0x00000F14         Thread ID: 0x00000624
Start Time: 10:33:12.625       Call Duration 00:00:00.406
ibsta: 0x2164       iberr: 0             ibcntl: 2(0x2)

6.  TestSRQ(0, FALSE (0))
Process ID: 0x00000F14         Thread ID: 0x00000624
Start Time: 10:33:13.062       Call Duration 00:00:00.000
ibsta: 0x164       iberr: 0             ibcntl: 2(0x2)

(6. repeated 300,000 times before MET/CAL gives up)

Do you have suggestions as to how to further troubleshoot this problem.

Sincerely
___________________________________________
Miikka Raninen
Technician
Accredited Calibration Laboratory
Finnish Airforces Materiel Command

0 Kudos
Message 1 of 3
(7,258 Views)

Looks like you expect SRQ generation immediately after Operation Complete.  Actually you get "1" from *ESR? query so the "operation" is actually complete.  To generate an SRQ from this event, you need unfilter the related mask registers in advance.

 

*ESE 1   -- unfilter OPC bit on Event Status Register

*SRE 32 -- unfilter ESB bit on Status Byte Register

 

Both commands are required. Did you send them before the operation?

0 Kudos
Message 2 of 3
(7,207 Views)
Solution
Accepted by topic author Ransu

Checked line 10 with a multimeter on the GPIB-cable. It was cut inside the connector and no one had marked the cable as such. Should've done this weeks ago. Smiley Wink

0 Kudos
Message 3 of 3
(7,159 Views)