IviScope Initialize.vi generates an 1074134995 error with TEKTRONIK TDS320 Oscilloscope

Error Message : 1074134995 IviScope initialize.vi
Driver status (0xBFFA002D) Entry missing from configuration file
My Configuration : WINDOWS NT 4.0, LABVIEW 6.02, MAX v2.1.3.14, NI488.2 v1.7, IVI engine v1.83,IVI Driver toolset 1.1, Instrument Driver v3.0, Oscilloscope TEKTRONIK TDS320,0,CF:91.1CT FV:v2.03
0 Kudos
Message 1 of 2
(3,206 Views)
Hello-

It sounds like this logical address is to be configured in MAX. Developer Zone has a document on how to do this in NI Developer Zone > Instrument Driver Network > Reference Materials > General. This document is called Using Measurement & Automation Explorer to Configure Your IVI System.

Randy Solomonson
Application Engineer
National Instruments
0 Kudos