Hi Team,

I am trying to get sweep measurement with Keithley 2460 source meter, with the help of instr. driver 2450 example. But it's not working!

so I just slightly modified the code, but it didn't work!.

I want to measure current with the source of voltage based on ON and OFF light.

I attached the sample code with this, can anyone please lookout this and let me what is the problem in the code.

please do the need full.

Thanks in advance

0 Kudos
Message 1 of 12
(4,139 Views)

The Keithley 2460 instrument has multiple programming languages.  Check which one is active with your system.  It appears that you want it in SCPI mode when running those VI's.

 

Check by performing the following:

1. Press the MENU key.
2. Under System, select Settings.

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
0 Kudos
Message 2 of 12
(4,111 Views)

Hi,

Yes, I checked, it's running in SCPI mode.

I doubted whether I made the wrong block diagram.

please do the needful.

 

0 Kudos
Message 3 of 12
(4,078 Views)
Solution
Accepted by kaust_priya31

Hi

Did you run the vi in debug mode following what was going on and eventually what the error is.

At first sight this looks ok, although several cases that are not needed. 

But I don't have a 2460 or alternative available to test this.

greetings from the Netherlands
Message 4 of 12
(4,004 Views)
Solution
Accepted by kaust_priya31

It looks to me like the code works just fine, but the graph isn't properly setup.

 

In the default example the graph has 2 y-axis, one for current one for voltage and they are very different scales.

 

In the attached modified code the graph has 1 y-axis for both current in the uA range and voltage from 0-1V.  But the data table clearly shows it is measuring.

 

I think the OP just needs to fix the plotting and its all good on the instrument side.

 

Craig

0 Kudos
Message 5 of 12
(3,985 Views)

Thank you so much for your reply.

0 Kudos
Message 6 of 12
(3,875 Views)

Thank you for your valid reply.

Yes, I run the code in debug mode, and the code works fine.

sorry for the delayed response.

0 Kudos
Message 7 of 12
(3,874 Views)

I am using Keithely2612B i try to measure piezoelectric voltage without any external source, voltage vs time sine wave any possible to measure if possible means can i vi of this instrument and i attached my vi of my i1.png2.pngnstrument

0 Kudos
Message 8 of 12
(986 Views)

This really doesn't track the solution from the above thread.  You may wish to post this as a new issue rather than a continuation.

 

If you are trying to read the piezoelectric voltage, then you need to set Source Output Configuration cluster to source current for the Source Function.  Otherwise, you will simply read the 1V input setting for the Voltage Level.

 

The plan would be to drive 0A and read/measure the voltage from the piezoelectric sensor.

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
0 Kudos
Message 9 of 12
(919 Views)

"Is it possible to fix the missing icon issue in the Keithley 6430 VI, and can a similar VI be created for the Keithley 2612B with equivalent functionality for measuring voltage, current, and resistance in my piezoelectric experiments?" and i attached the vi can you please change this one thank you

1.png