09-29-2006 03:44 PM
09-29-2006 09:38 PM
I am not going to claim to be an expert on impedance measurements, but I will try to help. The "open" and "short" compensations negate effects of your leads and test fixture. It essentially measures impedance and stray capacitance at different frequencies and applies a linearized offset to the measurements for your unit under test. You can perform the compensations with only your function generator in the circuit, and that should help. I no longer have acces to a 4194A manual, but Agilent had a number of tech notes for the 4194A, and impedance measurements in general, on their website. Quadtech also has some good tech notes on impedance measurements.
What kind of device are you testing? What frequencies are you using?
09-30-2006 05:29 PM
10-01-2006 05:52 PM
I have a better idea now, but I am still not real clear on what you are doing. I think that the analyzer is essentially a function generator. You could use the HP only, set the bias voltage to 2.5(it is AC), and either sweep or step through the 1 MHz. The way you have it, I think that to get good readings on the HP, you would have to sweep it at frequencies at least 2 (preferably more) times the FG frequency. Maybe you can explain why you chose 100k to 800k.
Brian
10-02-2006 01:49 PM
10-03-2006 07:03 AM
10-03-2006 10:24 AM
06-27-2008 04:27 PM
Has anyone know how to READ sweep frequencies (log mode)...or can point me to a Programming Manual?
thanks
06-30-2008 10:25 PM
06-30-2008 11:44 PM
Thanks,
From my research (including contacting Agilent) I assume that the Programming Manual doesn´t exist back then...
Also, the driver (from NI) doesn´t contain such a function (because it wasn´t included by HP as well).