Long measurements using Keithley 2450 SMU, USB Communication
My apologise for mark myself as person who solved my problem.
It was a mistake. Too much work.
Message 21 of 27
(1,793 Views)
Thanks a lot .Actaully my main problem is to make an IV curve and get the data from thr buffer. Do you have any IV curve program?
cheers
Message 22 of 27
(1,787 Views)
Download the Keithley driver from - https://www.tek.com/source-measure-units/2450-software-11
Under examples, look at "Sweep and Acquire Measurements.vi"
Craig
Message 23 of 27
(1,775 Views)
I tried to run Sweep and Acquire Measurements.vi but I can't. Your IV test works but I cannot run any of the other programs, I got errors about configurations of ranges
Message 24 of 27
(1,759 Views)
You should start a new thread since its obviously a new question.
What values are you trying to use?  The K2450 does have sourcing limits you need to be aware of .. https://www.tek.com/sites/default/files/media/media/resources/1KW-60904-0_2450_DataSheet_0.pdf
You can source up to 20V/1A, but above 20V the maximum source current is 100mA.  Make sure you set the compliance to 100mA if you try to sweep past 20V.  See the above spec sheet.
Cheers.
Message 25 of 27
(1,754 Views)
It doesn't matter what values limit level,delay or speed I use, It seems to be an autorange configuration issue
Message 26 of 27
(1,752 Views)
Start a new post and explain what you are trying to do explicitly.
Show the error messages you get, or the debug traces of the strings sent to the instrument and report actual settings.
Without more info we can't do much to help.
Message 27 of 27
(1,746 Views)