MAX reports error on scanning

It'll be the optical spec anny-  Yokogawa hasn't quite mastered GPIB

"Should be" isn't "Is" -Jay
0 Kudos
Message 11 of 15
(1,496 Views)

I was able to move the Coherent Power Meter to my USB network and get it off the GPIB network.  So if I can re-write software to point the new address, I should be fine.

Message 12 of 15
(1,496 Views)

Paul,

 

Sorry for the late reply, but your spy captures show exactly the same thing happening until line 3.  When asserting IFC (using ibsic), you can see that the return status does not have CIC (controller in charge) as true in the failure capture.  The passing capture correctly shows CIC is true in ibsta.

 

This is backed up by the fact that you get the ECIC error when using interactive control.  Out of curiosity, which command received this error?

 

Does one of these instruments have controller capability?  It could be possible that one of these is set to system controller which is causing problems in your setup.

 

I know that you have moved one of the offending instruments to USB, but I thought you may want another possible solution.

 

I hope this helps,

Steven T.

0 Kudos
Message 13 of 15
(1,471 Views)

Steven T wrote:

Paul,

 

Sorry for the late reply, but your spy captures show exactly the same thing happening until line 3.  When asserting IFC (using ibsic), you can see that the return status does not have CIC (controller in charge) as true in the failure capture.  The passing capture correctly shows CIC is true in ibsta.

 

This is backed up by the fact that you get the ECIC error when using interactive control.  Out of curiosity, which command received this error?

 

Does one of these instruments have controller capability?


Steven T.


The Textronix TDS scopes all have contontroller capability


"Should be" isn't "Is" -Jay
0 Kudos
Message 14 of 15
(1,465 Views)

I have added the GPIB-120B at the end of the GPIB daisy chain with the Coherent meter all by itself on one end and all my other instruments on the other.

 

Now when I scan for instruments, MAX reports "To many Listeners".

 

 

0 Kudos
Message 15 of 15
(1,417 Views)