Measurement and Automation Software does not find GPIB instrument
I've tried a different cable with the same result. Thanks for the program, I will try it
Message 6 of 14
(2,262 Views)
If you are using an Agilent (or HP) HP-IB card, the NI-MAX does not
show your HP-IB card on the device list even if the 'passport for
tulip' is properly installed for VISA operation. With the same reason,
instruments connected to an HP-IB will not be shown on the MAX software.
makoto
"Alison" wrote in message
news:5065000000080000005B380000-1007855737000@exchange.ni.com...
> I have installed a PCI-GPIB card and attached a HP scope. Initially
> the measurement and automation software found the scope and I could
> communicate with it in this. Also in Labview, I could control the
> scope but for some reason was unable to collect data from it. Now the
> measurement and automation software doesn't find the scope. It claims
> there are no instrume
nts attached to the GPIB. I have checked the
> cable is attached properly and windows control panel detects that the
> card is there and thinks it is working.
Message 12 of 14
(2,071 Views)
I've managed to get MAX to recognise the scope again, but the computer is still not getting waveforms from the scope.
Message 13 of 14
(2,071 Views)
Alison,
Would you be able to clarify a bit on what you mean by "computer not getting the waveforms"?  You mention using LabVIEW with your instrument driver, which commands are you using?
I'd also recommend the following steps:
1. select your GPIB port in MAX, scan for instrument, then select your device
2. open up a GPIB control panel (similar to VISA) or the GPIB Interactive Control interface from your start menu on your computer (under NI-488.2)
3. send a *IDN? query
you should recieve information about your instrument, confirming (on top of your device showing up in MAX) that your device is communicating properly.
Message 14 of 14
(1,692 Views)