I'm sure all of your problems are because of the two controllers connected to the same instrument. You can only have a single controller assigned as a master. From your description, you want to have the Agilent as the active system controller and the NI set as a talker/listener only. I don't recall where in MAX you do that but once done, before you can do a scan for instruments in MAX, the Agilent controller would have to pass control to the NI controller.
I'm not sure I even understand why you need a GPIB controller in PC2. There are a lot of easier (and cheaper) methods to communicate to a second pc. A simple cross-over network cable comes to mind.
04-29-2008 01:45 PM
04-29-2008 02:10 PM
Have you confirmed in the instrument manual that it actually supports the *IDN? command? Many older instruments do not and the message can be ignored.
Not sure about the addressing error. Have you tried setting the primary address for the controller to the default value of 0?