Problem with LabVIEW and Agilent E5810
I have an Agilent E5810 LAN/GPIB gateway and cannot get it working with
LabVIEW 7.0 under Windows 2000.
I have followed the instructions in this document:
http://digital.ni.com/public.nsf/websearch/E56D3D4866BE0A8A86256942006626F0/OpenDocument
which tell me to use Agilent VISA, rather than NI VISA.
The gateway works fine with Agilent VEE on the same computer. Whenever I
try to do a VISA Open I get this error message:
Error - 1073807342 occurred at VISA Open in measure_current.vi.
This error code is undefined. No one has provided a description for this
code, or you might have wired a number that is not an error code to the
error code input.
The VISA resource names I have tried are: (my instrument's GPIB address is
22)
TCPIP0::136.157.231.67::gpib0,22::INST
GPIB0::22::INST
TCPIP0::136.157.231.67::GPIB0:22::INST
but none works. The values TCPIP0 and GPIB0 come from Agilent's Connection
Expert software.
I would appreciate any help you can offer.
Richard Hill
Software Developer
Siemens Milltronics Process Instruments Inc.
Peterborough, ON
(705) 740 7017
(4,250 Views)
Please see the following link:
Try using "INSTR" instead of "INST" at the end of your address strings.
Logan S.
(4,233 Views)
What do you know, that did it!
I guess I couldn't see the forest for the trees ...
Thanks Logan.
(4,224 Views)