11-06-2007 10:15 AM
11-06-2007 10:43 AM
That error code indicates a framing error which indicates a mismatch of the com settings. Not sure why you mention timeout unless you have not recorded the error code correctly. The VI sets the com port to 1200 baud, RTS/CTS flow control, odd parity, 7 data bits, and 1 stop bit. Is this exactly how you setup Hyperterminal?
By the way, the VI was written by NI as far as I can tell. There's nothing on the download site that says it was authored by someone from Sartorius.
Also, the 1B50 is the hex equivalent of ESC-P.
11-06-2007 11:03 AM
11-06-2007 11:08 AM
11-06-2007 12:07 PM
11-06-2007 12:12 PM
11-06-2007 12:42 PM
11-06-2007 01:14 PM
I know it means framing error. That's what I said in my first response. It does not mean that the VI is reading before the instrument is sending. See the KB at http://digital.ni.com/public.nsf/websearch/13A20C0F9DF265FB86256FC60066D9C2?OpenDocument.
I did a quick search of the Sartorious web site but could not immediately find a manual for this model. Since I don't know how you changed the instrument's default com settings, I can't comment on that but maybe you should change it back to 1200, etc. and change the VI you modified to what it originally was and try it that way.
11-06-2007 01:49 PM
11-06-2007 02:20 PM