Hi,
I am trying to connect to the Agilent 4294A impedance analyzer through a LAN connection using a chosen IP address in order to control the device with a LabVIEW program but am receiving the following error from MAX:
Return status code: ocBFFF0011
Status name: VI_ERROR_RSRC_NFOUND
Insufficient location information or the device or resource is not present in the system.
This problem occurs very randomly. One second, the device is connected properly and I am able to run the LabVIEW program. Then, it disconnects and MAX gives me this error.
I do not think this is a problem with the LAN connection because the internet connection of my computer remains stable throughout.
If anyone has any thoughts please let me know.
Thanks,
Nika
08-25-2009 01:08 PM
Hi Nika,
Did you check the connection between the instrument and the hub?
Would you please try a different cable?
Regards,
Sammy
08-31-2009 10:36 AM
Hi,
I have tried changing the cable and again the problem arises randomly.
09-01-2009
05:39 PM
- last edited on
06-13-2024
02:10 PM
by
Content Cleaner
Hi Nika,
Let's update your NI-488.2 and NI-VISA drivers. You can find them at:
http://joule.ni.com/nidu/cds/fn/p/lang/en
Do you happen to have access to another 4294A to test with this system?
Please run an NI-SPY capture to see at what points there error occurs.
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000YHdrCAG&l=en-US
Regards,
Sammy Z.
09-02-2009 02:20 PM
Hi Sammy,
I have updated both drivers and still have the same problem and unfortunately I do not have other impedance analyzers to test the system with. However, I was not having this problem with the device a few months ago, which is weird.
I have attached the NI-SPY file for your reference.
Thanks for your help.
Regards,
Nika