:SENS1:CORR:COLL:THRU 1,2 don't work for E5071C
After the execution of :SENS1:CORR:COLL:THRU 1,2, the device does not execute this instruction as it should.
The following figure shows the correct effect of manually clicking the button.
(1,527 Views)
Is that a screen shot of the instrument?   Sending SCPI commands might not necessarily show as updated on the running software.  Some instruments require a completely different operating mode before they listen to SCPI commands.
Here's a few things you can try.
- Sendi the :SENS1:CORR:COLL:THRU 1,2 and then checking the status with the equivalent query.   :SENS1:CORR:COLL:THRU 1,2?  (not sure that's the exact format, just an example!!).
- Send other commands, like a PRESET or :READ?  to see if they have any effect on the instrument
Hope that helps.
Criag
(1,473 Views)
Accepted by topic author FairyXL
I found the solution on a sample document "E5071C to E5080B Code Migration.pdf", and after executing this command, then reversing the parameters to execute one side of this command, it works. Quite strange, but forget it.
:SENS1:CORR:COLL:THRU 1,2\n
:SENS1:CORR:COLL:THRU 2,1\n
(1,447 Views)