Setting up a GPIB communication with Keithley 220 and HP 34401A
Hi everyone,
I am using GPIB-USB-HS to connect both Keithley 220 Current Source and HP 34401A Multimeter and send commands via LabVIEW. I am using an existing old LabVIEW code, but I keep running into the problem (Error 7, file not found) when I send the command through GPIB Write Function, "INP:IMP:AUTO ON" (which sets the multimeter in high impedance).
And so, I am climbing down to the low level subVIs. However, it looks like I can't even run this simple voltage reader VI. There is simply no output string when I run the program, and so I assume I am not communicating with 34401A
Screen Shot 2018-06-21 at 3.47.00 PM.pngIn order to figure out whether the devices are connected to the computer via usb cable, I ran NI MAX program, and I GPIB-USB-HS "GPIB1" is visible under Devices and Interfaces.
What am I missing? One of my confusions is how to set GPIB address of two devices. Do I set them with the same address? I am confused then how, when sending a GPIB Write command, we can select one of the two devices (220 and 34401A).
(5,443 Views)
Hare you tried "Scan Instrument" in NI MAX?
GPIB Instrument Control Tutorial
After you finish the operation, you can see GPIB devices which are connected.
Certified LabVIEW Developer
GCentral
(5,416 Views)
Thank you. Scan for Instruments indeed show the two devices--Keithley 220 and HP34401A with GPIB address. This answers the question I had before.
I can now communicate with Keithley 220, but HP34401A is still non-responsive to LabVIEW commands. 34401A is in continuous measurement mode, as it shows correct voltage value on the front screen, as I control the current of 220 by LabVIEW.
Considering the computer recognizes 34401A, at least according to NI MAX, I believe there is some kind of a "button" I need to press on 34401A when I re-connect the GPIB cable. Someone else has been using this exact system on the same computer a week ago, and it was working just fine. I just disconnected GPIB cable and re-connected it, and it suddenly stops working.
(5,409 Views)
Accepted by appleSwell
it seems that you should change the instrument mode to "remote".
See the manual.
GCentral
(5,406 Views)
Other than placing the HP 34401A into remote, you will need to manually set the language to SCPI.
Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
(5,385 Views)