Tektronix TDS 220 Oscope not found by MAX using USB GPIB HS

Max 5.4 will not identify my Tk TDS 220 for any GPIB address I set it to.

TDS 220 not found.PNG

 

THANKS for any help!

 

I've already tried everything from these links:

http://www.ni.com/download/ni-488.2-3.1.1/3786/en/

http://sine.ni.com/apps/utf8/niid_web_display.model_page?p_model_id=289

http://sine.ni.com/apps/utf8/niid_web_display.download_page?p_id_guid=047216EC20B66FABE0440003BA7CCD...

0 Kudos
Message 1 of 4
(4,893 Views)

Hi VISluth! 

 

Is your instrument set to allow external control? You can find information about this in the instrument manual.

 

Also, while looking into the TDS 220, I found the following webpage which could suggest that you have defective hardware:

 

http://www.tek.com/service/safety/tds210-tds220

 

You could also check for IP between the instrument and your computer.

Applications Engineer
National Instruments
0 Kudos
Message 2 of 4
(4,859 Views)

I have the same (similar) problem.

Have you found a solution?

-------------------------------------------------------------------
Eugen Wiebe
Bernstein AG
CLAD - Certified LabView Associate Developer
0 Kudos
Message 3 of 4
(3,669 Views)

Solved for me!

I had to use a GPIB cable (because the USB->GPUB Adapter from NI was too big and did not correct plug)

and I also installed TekVISA (but did not replace the NI Visa).

And now it works fine for me.

Also the both drivers for LabView are very good!

Many thanks to NI!

 

-------------------------------------------------------------------