05-07-2009 02:16 PM
Solved! Go to Solution.
05-08-2009 10:24 AM - edited 05-08-2009 10:26 AM
Hello rcarr,
I am glad to see that you are using the LabVIEW Plug and Play (project-style) Instrument Driver for your instrument. Thank you for posting your VI and a screenshot of the error. It appears that you might be using the "Actn Start Stop Cont" VI incorrectly.
Please take a look at the examples that are included with the instrument driver and see how they are implemented. Also, please turn on Context Help (ctrl-H) and hover over the VIs to see more information about how they should be used.
Thanks,
NathanT
05-08-2009 02:08 PM
05-11-2009 09:06 AM
Hello rcarr,
The timeout error occuring in the Wait for RQS seems to show that the instrument driver is waiting for the instrument to send a service request (and that it is not being sent in the allotted amount of time). Thank you for attaching your VI. I think that you should try enabling the "Bus Trig Or Arm" control in the "Actn Start Stop Cont" VI. If that does not take care of the issue, I recommend looking through the instrument's manual and looking for the correct order of operations for a swept sine measurement.
Cheers,
NathanT