Unable to connect TDS210 oscilloscope using NI 488.2 GPIB-USB connector

I am using an NI USB-GPIB connector to connect to a TDS210 oscilloscope. I installed the NI 488.2 driver version 2.2.7 from the NI website. I am able to see the the NI GPIB interfaces using the Device Manager (see image below). Also, when I launch NI-MAX, I see the device listed with the following properties (see image below). I have connected a TDS210 oscilloscope to the GPIB controller. However, when I click the "Scan for Instruments" button, I get the following message:

"Instruments not found". The LED corresponding to READY on the USB-GPIB Controller is ON. However, the LED corresponding to ACTIVE is OFF.

 

I was able to connect to this instrument successfully using the Agilent USB-GPIB controller. I currently have only the NI controller connected to my system.

 

Any help would be  appreciated.

 

img1.jpg

 

 

img2.jpg

0 Kudos
Message 1 of 6
(4,495 Views)
Just a correction in my earlier post. I have installed the NI 488.2 driver version 2.7.1 that I obtained from the following link:
http://zone.ni.com/devzone/cda/tut/p/id/5326

Regards
Dinesh
0 Kudos
Message 2 of 6
(4,481 Views)
Can you clarify one thing? Before you installed the NI controller, did you completely uninstall all of the Agilent drivers?
0 Kudos
Message 3 of 6
(4,479 Views)

Hello Dennis,

I did not uninstall any of the Agilent Drivers from my system. I assumed that both can co-exist on one computer system.

 

Dinesh

0 Kudos
Message 4 of 6
(4,463 Views)
I don't think that is a good assumption. You can have Agilent VISA and NI VISA but if I remember correctly, there is a conflict with the GPIB drivers.
0 Kudos
Message 5 of 6
(4,458 Views)

Instead of using a Tektronix TDS210 oscilloscope, I tried using an NI Instrument Simulator hardware which has a GPIB interface. I was able to connect to this successfully. This seems to indicate that Agilent and NI drivers can co-exist on the same system (no problems thus far!). I think the issue might be the physical connection between the TDS210 and the USB-GPIB converter. I will look into this. Thanks for all your help!

 

Dinesh

0 Kudos
Message 6 of 6