I am stumped on this item. I have two PXI systems with built in GPIB controllers. One of the systems randomally started showing behavior where the double precision read commands returned gibberish data. On the other system, I am using the exact same code and everything works. Standard String commands work just fine (set voltage, set current, query set voltage etc.). I have tried the following:
Reset computer
Swapped Keithly instruments (the problem follows the PC)
Swapped GPIB cables
Updated 488.2 driver
Used seperate HS-USB to GPIB controller.
I will be reinstalling NI-VISA as well.
The only difference is that I installed some other LeCroy instrument drivers to the failing system. But the problem didn't occcur after I installed the drivers, it occured while my code was running.
Has anybody ever seen this before? What could other causes be?
Solved! Go to Solution.
12-12-2011 03:23 PM
12-12-2011 04:05 PM
Thanks, I will look into it. The real problem is that I have another identical system (initially) where everything is running just fine. And it has worked for 2 years previously.
03-27-2012 03:00 PM
The problem was with the Keithley, it was changing the format of the data without my knowlege.