Voltage and time control
Hello everyone~
I am using the labview and the NI GPIB-USB-HS card to control the Agilent N8736 DC power supply.
The goal :I want to output the 24V voltage,then drop to 9V ,then rise to 24V ,the fall time and rise time shall be less than 10ms.
The question: how to assure the 10ms time?
I have download the Agilent N8736 GPIB driver from the NI web, and based on the model, I write the follow program listed in attach, and the test result is shown also.
I cannot assure the 10ms time, the actual test time was more than 300ms.
Thanks a lot~~
(5,008 Views)
I would say: You can't achieve the <10ms  with that powersupply. If you look into the specs (http://cp.literature.agilent.com/litweb/pdf/5990-3881EN.pdf) there are "output response times" and "command response times" much higher than your goal.
(4,995 Views)
Accepted by yishui
You may be able to get what you want with an extra power supply, a solid state relay, and a DIO that is hardware timed.
(4,989 Views)
mkossmann:
           Thankyou very much. I have read the instructions before, but cannot understand the output response time and command response time clearly. Now I see it, Thanks a lot~~~
(4,967 Views)
crossrulz:
     Thanks a lot. I will  follow your advices and try other methods to attach my goal.
     best to you~~
(4,966 Views)
