What I write to buffer is still in buffer when I read
I'm trying to communicate with an instrument via serial. According to the instrument's manual, when I write "SP", the instrument should respond with with "Maximum Speed: [value]". When I try this in NI MAX, however, what I read is "SP \r \r Maximum Speed: [value]". It seems the command I write is still in the buffer when I read from the buffer. Is this typical for serial communication? Is there a way to avoid this? I can't clear the buffer after I write "SP" because the instrument responds immediately, so then I'd be clearing the information I'm trying to read.
(1,046 Views)
Accepted by jrdwight
It may be typical for some instruments to echo the command along with the response. Check your instrument manual if that is expected.
Santhosh
Soliton Technologies
(1,033 Views)
Also, is this 2-wire RS-485?  I rarely deal with RS-485, but I remember reading what I sent before being able to read the response when I did.
You need to give a lot more information on your setup before we can really tell you if this is expected.  What is the instrument?  What is the protocol being used (RS-232, RS-422, RS-485, etc.)?
(987 Views)
I determined this was the expected behavior for the instrument I was communicating with (which was via RS-422 by the way). I had it in a mode where it echoed every command it received. I was able to change the operating mode so that this echo is disabled.
(976 Views)
That is common for instruments that want to interface with a terminal program such as Putty.  Thank you for letting us know what you found.
(965 Views)