Ahhh, very interesting. It sounds like you have two 488.2 drivers installed (except for the fact that both instruments don't respond with double characters even though they are connect to the same board, which is very strange)
I'd test if it happens when you use VISA and if NI-Spy shows the duplicate characters.
In any case, it's likely that something is indeed wrong with your installation. I'd bet that you have LabVIEW 6.0.2 or earlier, did a "complete" installation of LabVIEW, then later installed NI-488.2 separately.
Reinstallation Instructions:
*Remove All NI software through start -> settings -> control panel -> add/remove programs
*Open Windows Explorer and go to c:\program files, if there is a National instruments folder delete it and all of it's c
ontents
*Shut down the computer and remove the cards
*Install LabVIEW, do a custom install and cross out the NI-DAQ, VISA, and 488.2 drivers so they won't be installed
*Install 488.2 software
*Shut down and put in the GPIB card
All drivers can be found here