08-04-2009 05:06 AM
Hi Guys,
I want to build a GPIB structure which will have several controllers to controll several instruments. like the figure shown below. Is it possable?
08-04-2009 01:36 PM
ghlaah,
No you can't do want you have shown above. You can only have 1 GPIB controller-in-charge at a time on a GPIB bus. You want to be able to communicate from Client 12 with the SG at the same time as Client 14 communciates with the DMM - not possible.
To accomplish this you can either make sure all the instruments have LAN ports and just have a totally Ethernet system or alternatively you could replace the GPIB Hub, you show, with an Ethernet -to-GPIB controller. All the Clients hook up to the Ethernet-to-GPIB controller via Ethernet and all the instruments connect to the GPIB side. This should work according to NI's data sheet for the NI GPIB-ENET/100.
08-04-2009 06:21 PM
Another option besides the GPIB-ENET/100 controller, is to connect all the instruments to a GPIB interface on one of your computers, then use NI-VISA Server to allow remote access to those instruments over ethernet.