08-03-2011 10:35 AM
Solved! Go to Solution.
08-03-2011 10:48 AM
Is your PCI-GPIB card made by NI? If it is open MAX (measurement and automation explorer) and expand devices and interfaces on the left hand side. You should see your PCI GPIB card there usually called GPIB0. Click that and on the right hand top menu click "Scan for instruments". If your agilent is powered up and responds to 488.2 IDN? commands it should appear at the bottom at the PAD address listed.
08-03-2011 12:24 PM
08-03-2011 12:47 PM - edited 08-03-2011 12:48 PM
That's just wrong. You do NOT set the address of the GPIB controller to the same address of any instrument that you have connected. You should have just left it alone at the default address of 0. I would suggest searching for a GPIB tutorial. Each device on the bus must have a unique address.