2 parallel data acquisition loops

Hi. I use 6363 OEM daq and I would like to acquire 2 digital channels at different sampling rates ( one 2MS/s and 10kS/s)

So I tried to use 2 parallel data acquisition loop but when I run one of them works fine but the other one gives me this error and stops

 

Error -50103 occurred at DAQmx Start Task.vi:2

Possible reason(s):

NI Platform Services: The specified resource is reserved. The operation could not be completed as specified.

 

Could you please let me know how can I acquire bothe signals at differnt sampling rate with an example

Thanks

0 Kudos
Message 1 of 5
(3,486 Views)
You can't. This error is common and has been discussed numerous times. Since you have a single clock for the one digital resource, you have to acquire all channels in a single task and at the same rate.
0 Kudos
Message 2 of 5
(3,479 Views)

Thank you Denis. I have to acquire 6 digital signals. The sampling rate of 5 of them can be as low as 1000S/s but one of them should be atleast 1MS/s and then I need to log digital waveforms into a TDMS file. The problem I have is if I use 1MS/s for all of them then I make a very big TDMS file while 80 per cent of the samples are really not needed . I finally want to read the data from TDMS and graph the waveform and as I said I will waste a lot of memory if I use 1MS/s for the other 5 signals. Can you help me on this. How can I optimize this process?

Thanks

 

 

 

 

 

 

 

 

 

 

  

0 Kudos
Message 3 of 5
(3,440 Views)

You'll have to collect them all at the highest sampling rate you need and then downsample the other channels in software before you save them.  If you sample at 1MS/s and only need 1kS/s just pull out every thousandth element into a new array and update the waveform dT (1000 times larger).  I think there's a built-in Waveform processing function that will do both steps in one function, but I'm not looking a copy of LV at the moment.  It's not particularly hard to code manually if I'm wrong.

-Ian Konen
0 Kudos
Message 4 of 5
(3,420 Views)

Yes, there's a downsample VI, if it just decimates the array or average the parts i dont know, though it can be rather easily tested. If downsampling i feel it'd be good if you could average the downsampled part. Assuming alot of redundant samples it might be moot as differences would be small, but in e.g. a 2:1 or 3:1 downsample with frequencies nearing the Nyquist(?) frequency it would do some difference.