03-10-2010 03:50 PM
I am trying to configure pulse mode on Agilent 33120A using RS-232. I finally figured that the relevant vis on instrument driver does NOT support 33120A!!!
Have anyone got any ideas?
Solved! Go to Solution.
03-10-2010 04:03 PM
Does the instrument support pulse generation? It's not listed on the data sheet like it is for the 33210, etc..
03-10-2010 04:07 PM
Dennis Knutson wrote:Does the instrument support pulse generation? It's not listed on the data sheet like it is for the 33210, etc..
I think the OP means "burst" or a PWM square wave. 33120A is a full featured AWG. I've gotten tons of "pulses" out of 'em
03-10-2010 04:12 PM
I'm using it to generate a single pulse by using the square wave function in burst mode. Here are the commands that I send:
*RST; *WAI;*ESE 0; *SRE 0; *CLS; *WAI;
*PSC 0; *ESE 188; *SRE 32;VOLT:OFFS 0;
VOLT MIN;OUTP:LOAD INF;VOLT:UNIT VPP;
FUNC:SHAP SQUARE;TRIG:SOUR BUS;
BM:PHAS 0;BM:NCYC 1;BM:SOUR INT;
BM:STATE ON;PULS:DCYC 50;FREQ 6.25;
I then send a software trigger.
As Jeff said you can also use the ARB function (you then have to create the waveform shape that you want.
Ben64
03-10-2010 04:15 PM
Yes, 33120A supports it. But, the vis in the 33xxx driver specifically mention that they can NOT be used with 33120A.
I need to configure the device as follows.
Freq = 1000 Hz
Amp = 1.5 V
Offset = 750 mV
Burst CNT = 5 cycles
Burst Rate = 100
Burst Phase = 0
Burst Source = Internal
03-10-2010 04:18 PM
03-10-2010 04:24 PM
ben64 wrote:Don't know why some smileys were inserted in the text.
So here is a screenshot (sorry the the large size).
Ben64
Don't worry. Right now you cannot unescape smilies on the editor. Laura F. is still working on forum upgrades. How are they doing Laura?
03-10-2010 04:28 PM
spider wrote:Yes, 33120A supports it. But, the vis in the 33xxx driver specifically mention that they can NOT be used with 33120A.
I need to configure the device as follows.
Freq = 1000 Hz
Amp = 1.5 V
Offset = 750 mV
Burst CNT = 5 cycles
Burst Rate = 100
Burst Phase = 0
Burst Source = Internal
the 33120A ishas been "Obsoleted" for about 5 years by Agilient. Therefore newer agilient drivers will not "support" them. Yet, the 33120A is fully SCPI compliant so any AWG driver that uses SCPI MLM's will need only minor tweeks to support the 33120A command set. Esentially you need to tweek range checking and the model name.
03-10-2010 04:29 PM
It should be pretty straightforward to add the burst function to the driver. No driver is 100% complete and as the download page states, it was tested with the 33210 and 33220.