04-29-2009 04:23 PM
I am using LV 8.2.1. Serial communication @ 9600 and using some drivers I got from the NI website. My code does not throw any errors but the instrument does for each time I run the VI???
I have tried delays and configuring triggers, changing the baud rate and nothing is working.
I am using a 34901A card for temperature sensing and only reading one of the many probes it has, does anyone have any advice for me?
04-29-2009 04:25 PM
04-29-2009 05:10 PM
My experience with the 34970A and 4-wire PRT temp sensors was that I got an error using the default setting for a PRT but when I set the actual type, the error went away. I'm assuming that you get an error (red light) on the front panel of the 34970A and Labview doesn't report anything.
Stuart
04-29-2009 06:45 PM
04-30-2009 09:40 AM
I finally found the code. "+111, "Channel list: slot number out of range"
I have it in slot 1 and am measuring channel 101. Both of these values I have wired to the drivers. I am sure this would easily be fixed if I had the manual but I do not. If anyone can link me to a copy I would appreciate it. I came up empty on agilents website.
Thanks for the replies so far.
04-30-2009 09:44 AM
04-30-2009 09:45 AM
I am sure this would easily be fixed if I had the manual but I do not. If anyone can link me to a copy I would appreciate it. I came up empty on agilents website.
34970A manuals can be found here.
04-30-2009 10:25 AM
04-30-2009 11:35 AM
Try MEAS:TEMP? TC,J, (@101)
Note the , before the (
-AK2DM